Re: Spectrophotometer's Illuminant and Angle Affect



Dear Mr. Whitfield,

On Jul 17, 2006, at 7:27 PM, Bill Whitfield wrote:

Greetings,

I have noticed that the illuminate and angle for my i1 and DTP41 are D50 and 2 degrees. I am guessing they are set the same so their readings will be close.

The settings on the production floor's spectrophotometers (DataColor) are D65, 2 degrees for our local plant and D50 10 degrees for our corporate plant.

Readings for the same PMs color (186) on the local production floor spectrophotometer is different than the reading for the same color on the corporate spectophotometer. I would imagine is because their illuminate and angles are different(?). Does the illuminate and angle make a difference, unless one is trying to standardize device readings?

These are not the instrument's measurement geometry and illuminant, but the measure of the field of view for the standard observer and the reference illuminant by which the instrument's spectral data will be processed. The DTP-41 and the Eye-One both use tungsten-halogen illumination (roughly corresponding to CIE standard Illuminant A) and 45/0 measuring geometries for reflectance measurements. The "45" in the term "45/0" refers to the angle of the illumination from the normal to the measurement surface. The "0" in the "45/0" term is the position of the instrument's optical pickup from the measurement surface normal. Simpler put, the instrument is looking straight down on the spot illuminated by the instrument's lamps set to a 45 degree angle.


I do not know what type of illumination the DataColor is using, but most DataColors use an integrated sphere for the measuring geometry. This is a reflective sphere with a hole on one side for the sample, another hole approximately 10 degrees down from the normal on the opposite side and a hole down the side, and out of view, for the illumination. The sample is lit from the reflected light from all directions, simulating a more real-world view with light coming from all directions to a surface. Because there is a specular reflection visible to the optical pickup from it's 10 degree from normal position, most integrating spheres have a black trap that can be positioned at the opposing 10 degree position. This has the effect of absorbing the reflection to give a "specular excluded" measurement, or, by placing a reflective material at this position, producing a "specular included" measurement. The "specular excluded" measurement is roughly equivalent to a "45/0" geometry measurement.

Integrating spheres are the recommended method for measuring fabrics, rough papers and other textured surfaces. "45/0" geometries are recommended for smooth surfaces.

Once the sample is measured spectrally, the illuminant and sensor components are subtracted from the measured spectrum to give the reflectance of the sample itself. Once the spectral data from the sample is obtained, the spectrum can be recalculated with any illuminant and observer. There are two standard observers; 2 degree and 10 degree. The 2 degree is used for small samples that occupy an area up to 4 degrees of arc, the 10 degree observer is used for samples greater than 4 degrees of arc in area.

The spectral data from the DTP-41, Eye-One and DataColor instruments may be compared, but the colorimetric data from D50/2-degree cannot be compared with D65/2-degree or D65/10-degree.

The data for your PMS patches should be different for instruments with differing observer and illuminant calculations. Also since there have been many generations of PMS color selectors, unless you are measuring the same patch from the same sample book, you are almost guaranteed to get different colorimetric values.

If you are interested in learning more about this subject, I can recommend "Billmeyer and Saltzmann's Principles of Color Technology" by Roy S. Berns as a good beginning to intermediate text.

Regards,

Robin Myers

_______________________________________________
Do not post admin requests to the list. They will be ignored.
Colorsync-users mailing list      (email@hidden)
Help/Unsubscribe/Update your Subscription:
http://lists.apple.com/mailman/options/colorsync-users/email@hidden

This email sent to email@hidden
References: 
 >Spectrophotometer's Illuminant and Angle Affect (From: Bill Whitfield <email@hidden>)



Visit the Apple Store online or at retail locations.
1-800-MY-APPLE

Contact Apple | Terms of Use | Privacy Policy

Copyright © 2007 Apple Inc. All rights reserved.