Re: Monitor Calibration
[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

Re: Monitor Calibration


  • Subject: Re: Monitor Calibration
  • From: "J.Raimar Kuhnen-Burger" <email@hidden>
  • Date: Wed, 18 Nov 2009 18:25:51 +0100
  • Thread-topic: Monitor Calibration

Hallo Klaus,

Keep in mind that these have just only be measurements of the white.
Especially for wide gamut units, all colorimeters need a correction. And, as
I´ve wrote, they also need a correction for every other display that does
not match the device manufacturer´s internal standard.

With precise corrections everything is getting better. The Delta between a
DTP94 and a CS2000 drops to 0.001uv for the whitepoint.

Using EDID data is not the right way as the EDID data is in most cases not
identical to the real values. As soon as you change the whitepoint or adjust
the LUT, the xy chromaticity of RGB change and that is not compensated by
the EDID even if it is precise.

Best regards

Raimar


am 18.11.09 15:41 schrieb Klaus Karcher unter email@hidden:

> Hello Raimar,
>
> thank you very much for this valuable information.
>
> J.Raimar Kuhnen-Burger wrote:
>> Hi Graeme,
>>
>> i´ve showed inter instrument agreement only between the EyeOnes, Spyder3s
>> and DTP94s, not against the Minoltas. The Minoltas are used for comparison
>> and for reference only. But here are some simplified numbers - averaged from
>> 10 devices that are quite close to each other - against the CS2000 for your
>> records. To trace the real behavior, there have been no device corrections
>> used.
>>
>
> I've normalized the measurements to L=100 and calculated DE00 to the
> reference instrument to make your results more vivid for me:
>
>> sips72ccfl
>       ID nits XYZ_X XYZ_Y XYZ_Z     DE00
> 1   ref  121   118   121   105 0.000000
> 2 dtp94  115   110   115    94 6.158537
> 3  spy3  117   113   117    91 7.303911
> 4  i1d2  114   112   114    93 4.838307
> 5   i1p  118   113   118    98 5.134003
>>
>> spva92ccfl
>       ID nits XYZ_X XYZ_Y XYZ_Z     DE00
> 1   ref  124   121   124   110 0.000000
> 2 dtp94  130   130   130   107 6.969239
> 3  spy3  134   132   134   122 4.250049
> 4  i1d2  130   126   129   107 4.341224
> 5   i1p  130   126   130   111 3.200485
>>
>> sips102ccfl
>       ID nits XYZ_X XYZ_Y XYZ_Z      DE00
> 1   ref  120   114   120   102  0.000000
> 2 dtp94  117   115   117    88 10.332352
> 3  spy3  115   109   115    95  2.636599
> 4  i1d2  115   116   115    86 14.296324
> 5   i1p  121   116   121   100  2.646035
>>
>> sips108led
>       ID nits XYZ_X XYZ_Y XYZ_Z     DE00
> 1   ref  122   117   122   102 0.000000
> 2 dtp94  120   119   120    93 8.789449
> 3  spy3  117   112   117    98 1.892420
> 4  i1d2  115   114   115    93 8.141067
> 5   i1p  119   115   119    97 2.704125
>
> I think some of the values are rather alarming. In consideration of
> these results, I ask myself if it might be even better to count on the
> manufacturer's EDID data then on own measurements with inaccurate and
> unsuitable devices in some cases.
>
> Klaus

_____________________________
J.Raimar Kuhnen-Burger

Master of Art (Farbkommunikation)
Diplom Designer (Farbe)
Kunst- und Medienwissenschaftler

Milchberg 26
D-31177 Harsum

Tel./Fax.05127-931006
www.concolori.de

Skype: jrkuhnen
iChat: jrkuhnen


 _______________________________________________
Do not post admin requests to the list. They will be ignored.
Colorsync-users mailing list      (email@hidden)
Help/Unsubscribe/Update your Subscription:

This email sent to email@hidden

  • Follow-Ups:
    • Re: Monitor Calibration
      • From: Klaus Karcher <email@hidden>
References: 
 >Re: Monitor Calibration (From: Klaus Karcher <email@hidden>)

  • Prev by Date: Re: Monitor Calibration
  • Next by Date: Re: Monitor Calibration
  • Previous by thread: Re: Monitor Calibration
  • Next by thread: Re: Monitor Calibration
  • Index(es):
    • Date
    • Thread