Ethan Hansen wrote:
> The normal failure signature for a Spectrolino is the lamp output dimming. This is seen
> in measurements by darker colors having lower than actual L* values. Once the lamp
> starts dimming, the Spectrolino is on borrowed time. Fairly frequent re-calibrations
> are required - figure on getting 50 to 100 measurements of ~2K patches between cal
> cycles.
> We also haven't found a source
> for replacement lamps. We cannibalize a stockpile of instruments to keep the current
> ones running.
Probably not much help, but the lamps seem to be 2.5V Medical lamps, possibly
Quarz Halogen, given their higher color temperature (> 3000K) and the desirability
for longevity.
Something like <https://www.precisemedical.com.au/collections/medical-lamps-and-bulbs?page=1>
Given the lamps are glued into place, it wouldn't be easy replacing them though.
If they (and the similar lamps in the i1Pro) are Quartz Halogen, then might
explain some odd characteristics I've noticed about them...
Graeme Gill.