Re: instrument calibration
Re: instrument calibration
- Subject: Re: instrument calibration
- From: Roger Breton <email@hidden>
- Date: Sun, 30 Jun 2002 11:49:01 -0400
>
I have a white Spectralon reference tile that is calibrated to NIST standards
>
that I can
>
measure spectrophotometers against. When your spectrophotometer gets too far
>
from the
>
standard, it should go back to the manufacturer for recalibration and
>
recertification. I
>
know that GretagMacbeth and X-Rite have services for recalibrating their
>
instruments. You
>
>
should check with SpectroStar to see if they also offer this service. If you
>
really want
>
to check the accuracy of the spectrophotometer, erbium oxide and holmium oxide
>
standards
>
can be used. They have much more precisely defined peaks and valleys in their
>
reflectance
>
>
spectra that can be used for determining wavelength accuracy of your
>
instrument.
>
>
Robin Myers
The problem with White reference standard, IMHO, is that it won't catch
wavelength scale errors. I used to own a Spectrocam for a while, until I
returned it for refund, and I remember checking its response on a similar
White calibration standard, one that came with a Datacolor Spectroflash
sphere instrument. Well, as long as I kept measuring white and neutral
colors, there was always very close agreement between the Spectrocam and the
Datacolor, and a 938 and a Spectrolino: very little difference indeed
because none of these measurement involved the wavelength scale.
But the moment I tried to measure saturated colors, like the RGBCMY patches
on a ColoChecker, all bets were off.
Regards,
Roger Breton
Laval, Canada
email@hidden