Richard,

You should not compare data taken from two different instruments,

The sentence before this one in my response was:

"The Eye-One is a 45-degrees/0-degrees instrument, compared to your two sphere instruments, and assuming specular-included (SPIN) measurements, the L* values should be slighly lower for glossy samples in the Eye-One case. You should not compare data taken from two different instruments,..."

I should have added the word "geometry" after "two different instruments" to make it clearer; my mistake. You can easily compare data from two instruments of the same geometry, as your customer asks, but not the Eye-One vs a sphere instrument.

I believe the machine offers 20nm spectral resolution.

And "reported" resolution versus actually measured resolution may be different. This is just another aspect that complicates inter-instruments comparison.


They both have SPIN.

I saw that some recent Datacolor model 600 spectros have both SPIN/SPEX options.


At this point, confirming fluorescence with a simple spectrum check with an Eye-One with NO UV-cut seems the way to go.

Regards,

Danny

email@hidden
www.BabelColor.com

Could the fluorescence also uncover the use of brighteners?
I wouldn't think they would do that but could brighteners cause such a difference?

Yes, this is how fluorescence affects the color.
The Optical Brightening Agent (OBA), either in the substrate or the dye/pigment, will do that.


On Sat, 26 Aug 2006 08:14:19 -0500
 Richard Brackin <email@hidden> wrote:

On Aug 25, 2006, at 5:26 PM, <email@hidden> wrote:


Richard,

You should not compare data taken from two different instruments, as you mention in your point 2. Also, Eye-One Share uses a 2 degrees Observer and your other data was computed with the 10 degrees Observer; not to be done also.
That is effectively what the customer is having us do.
ColorGuide Sphere 6830 readings vs. DataColor 600 readings.



Your point 3, about various resolution for different instruments goes in line with my comment (in my first response) on the sharp cut-off often seen on (bright) yellow, orange and red colors.
Yes. I definitely took note of that.
I believe the machine offers 20nm spectral resolution.


Look at the spectrum with the Eye-One. If it goes over 100 at some wavelengths, around 590 nm, you have your answer on fluorescence. (Note: You can also use the BabelColor software in demo mode for this measurement ;-) )
We'll take a look at your software.
Could the fluorescence also uncover the use of brighteners?
I wouldn't think they would do that but could brighteners cause such a difference?



Is your Eye-One a UV-cut model?
The one I'm using is not UV-cut.
We have an eye-one that is a UV-cut model at our other facility to calibrate/profile our UV presses over there.


Does one or both of your sphere instruments have a specular-included (SPIN)/specular-excluded(SPEX) option?
They both have SPIN. I'll have to dig around the menus and see if they have SPEX but I think they're SPIN.
It's a BYK-Gardner ColorGuide Sphere SPIN model CG-6830 (I should have initially mentioned they were SPIN).
The two devices we have are pretty much identical.
.

Danny

email@hidden
www.BabelColor.com



Thanks for your time and information. Richard

_______________________________________________
Do not post admin requests to the list. They will be ignored.
Colorsync-users mailing list      (email@hidden)
Help/Unsubscribe/Update your Subscription:
This email sent to email@hidden


References: 
 >Question: Measuring Delta E from two different spectrophotometers (odd problem) (From: Richard Brackin <email@hidden>)
 >Re: Question: Measuring Delta E from two different spectrophotometers (odd problem) (From: <email@hidden>)
 >Re: Question: Measuring Delta E from two different spectrophotometers (From: Richard Brackin <email@hidden>)
 >Re: Question: Measuring Delta E from two different spectrophotometers (From: <email@hidden>)
 >Re: Measuring Delta E from two different spectrophotometers (From: Richard Brackin <email@hidden>)
 >Re: Measuring Delta E from two different spectrophotometers (From: <email@hidden>)
 >Re: Measuring Delta E from two different spectrophotometers (From: Richard Brackin <email@hidden>)

  • Prev by Date: Re: Measuring Delta E from two different spectrophotometers
  • Next by Date: Monaco Optix v3.0?
  • Previous by thread: Re: Measuring Delta E from two different spectrophotometers
  • Next by thread: ProPhoto and ROMM
  • Index(es):
    • Date
    • Thread