Re: Quoting accuracy of color measurement instruments
Re: Quoting accuracy of color measurement instruments
- Subject: Re: Quoting accuracy of color measurement instruments
- From: Marco Ugolini <email@hidden>
- Date: Wed, 31 Oct 2007 09:13:50 -0700 (GMT-07:00)
Graeme Gill wrote:
>No, the underlying CCD measures (approximately) at 3.5 nm spacing (and
>naturally these bands overlap), and then it is re-sampled (and applies
>wavelength calibration) down to 10nm spacing.
Graeme,
I'm having a hard time understanding why, since the measurements are taken at a higher resolution, they are then resampled to a significantly lower resolution. Why not leave them at the original resolution? Or resample them to 5nm instead? Are there any advantages in resampling to 10nm instead?
Regards.
Marco Ugolini