On 04/23/2011 07:39 PM, Robin Myers wrote:
By "UV enabled" I believe Mr. Dinkla was referring to an instrument
without a UV cut filter, so it is enabling measurement of UV
wavelengths. In which case he correct in his assertion that it
measures a wider spectral range.
The UV Cut filter employed in the i1Pro removes all illumination
below 400 nm. Without illumination there should be no reported data
in this region. So the actual range for the i1Pro UV (this is
X-Rite's nomenclature for an i1Pro with a UV-cut filter installed) is
400-730 nm in reflectance measurement mode.
However, X-Rite, for some unknown reason, creates and reports
extrapolated data in the 380-400 nm region. At best this leads to
false interpolations of data in this region, at worst it creates
highly erroneous data which causes problems further along. I have
measured papers with FWA additives that sometimes cause their
interpolation algorithm to produce values in excess of 200%
reflectance in the 380-400 nm region.
I have a more thorough report on this issue in my Chromaxion
newsletter available at
http://rmimaging.com/information/Chromaxion_Issue_1.pdf.
To remove issues with bad extrapolation data in my SpectraShop
program, I set the data to 0% reflectance in this range for the i1Pro
UV and the Spectrolino when used with the UV Cut filter.
Robin Myers
Mr. Meyers described it correctly for the Eye 1 Pro. I use the term "UV
enabled" as in several cases the "UV-cut" models do not have any potential
to measure the shorter wavelengths so do not have to cut UV light out
with a filter. Calling the last "UV-cut" isn't correct, they could be
called "Non-UV" I guess. With about 4 categories it becomes difficult to
call one "Normal".
The ColorMunki in reflective mode measures down to 430 Nm
and copies the 430 Nm result in the data down to 380Nm.
I doubt the Spyder3 is any better on the short wavelengths.
The HP Z3100 etc is measuring with 20Nm intervals at the range of
400-700 Nm. Interpolating down to 10 or 5 Nm when selected in Color
Center measurements export.
I wonder what the "UV cut" mode of the iSis does on wavelength range in
reflective mode. The limitation is in the light source and most likely
that light source is similar to the Z3100 white LED. The optical
interval is 10 Nm.
The SpectroCam (UV enabled) used to measure from 380 to 750 Nm in 5 Nm
intervals.
The best UV enabled spectrometers measure a wider range in general. The
"UV cut" models cut at 400 Nm, the "Non-UV" at 400-430 Nm and their
limit varies between 700-730 Nm at the other end.
--
Met vriendelijke groeten, Ernst
Dinkla Gallery Canvas Wrap Actions
| Dinkla Grafische Techniek |
| www.pigment-print.com |
| ( unvollendet ) |
_______________________________________________
Do not post admin requests to the list. They will be ignored.
Colorsync-users mailing list (email@hidden)
Help/Unsubscribe/Update your Subscription:
This email sent to email@hidden