- From: Roger Breton <email@hidden>
- Date: Mon, 11 Aug 2014 09:21:54 -0400
Dear Graeme,
Allow me to ask the $1000 question : how comparable is your OB correction method is to "M1" measurements?
Best / Roger
www.graxx.ca
-----Original Message-----
From: colorsync-users-bounces+graxx=email@hidden [mailto:colorsync-users-bounces+graxx=email@hidden] On Behalf Of Graeme Gill
Sent: 11 août 2014 08:08
To: ColorSync
Subject: Re: X-Rite EyeOne iSis with M1 option
Roger Breton wrote:
> My skepticism with regards to software methods has to do with the
> little I know of tungsten halogen lamps : if the lamp emits little or
> nothing in the 360nm range, where OBs are typically excited, well?
It's not a matter of how little UV an incandescent lamp emits in absolute terms, it's a matter of how much it emits relatively, since spectrometers measure reflectance. i.e. the ratio of UV to blue light output of the lamp is quite sufficient to trigger obvious FWA/OBE fluorescence in typical papers, just not as much as a using a true D50 light source.
Graeme Gill.