Scott Martin wrote:
Perhaps I should have said that differently. The Spyder 3 doesn't have a
baffle to eliminate ambient light. When the Spyder3's suction cup is
Maybe that's another problem as well. The problem I've seen
is that the the dark readings are pretty inconsistent. For instance,
if I try and measure my CRT's black several times I
get the following XYZ numbers:
0.050 0.009 0.299
-0.037 0.060 0.048
0.009 0.005 0.113
-0.035 0.057 0.044
By contrast, the Spyder 2 gets:
0.098 0.077 0.061
0.097 0.077 0.061
0.097 0.076 0.063
0.102 0.077 0.067
Not too bad.
[The above is using the manufacturers supplied drivers
for these instruments. Stepping outside this and using
longer integration times for the Spyder 3 improves its
situation.]
Given the typically poor blacks on LCD's, maybe this isn't
an issue for the Spyder 3 for most users, although the introduction
of LCD displays using dynamic LED backlighting ("BrightSide"
HDR technology), may disturb that comfort.
Graeme Gill.